Postby jlopez » Mon Nov 09, 2009 5:50 pm

Many times I have had to clean a single crystal that I am not familiar with and have found that every person that works with that type of crystals uses a different "recipe" to get the sample clean. I thought that this would be a good place to make a collection of the recipes and then they would be accessible to everybody when we need to have an idea of the cleaning procedure for a sample that we are not used to. Said that, the question for everybody in the community would be, what sample you use and which way do you clean it?
jlopez
 
Posts: 8
Joined: Sat Oct 17, 2009 3:31 pm

Crystal Cleanning: Au(111)

Postby jorgeab2 » Wed Nov 11, 2009 1:41 pm

For cleaning a Au(111) single crystal. Bombard with Ar+ with a current of ~2 uAmps and a voltage of 500V for about 30mins. Then anneal to ~620C for 5 mins followed by 30 mins at ~420C. Repeat the same until the sample is clean. I check that is clean by getting the herring-bone reconstruction by STM.
User avatar
jorgeab2
 
Posts: 9
Joined: Thu Oct 15, 2009 8:14 am

Crystal Cleanning: Cu(111)

Postby anibalbosco » Mon Nov 16, 2009 10:37 am

Cu(111):
Bombard with Ar+, 2uAmp, 1kV, 30mins. Anneal to 25 mV (type k thermocouple) for 30mins to flatten the bombarded surface. Check cleanness by Auger. If not completely clean,repeat cycle.
anibalbosco
Site Admin
 
Posts: 8
Joined: Sat Sep 12, 2009 1:33 pm
Location: Milwaukee, Wisconsin

Crystal Cleanning: Pd(111)

Postby jorgeab2 » Thu Jun 10, 2010 1:02 pm

Pd(111)
1.Ar+ sputtering 1kV, 4uAmp, 30mins.
2.Anneal to 1000K for 5 mins.
3.Anneal between 740K and 875K while dosing O2 on the surface (4E-8Torr seems a reasonable pressure) for 30 mins moving in this Temp range during this time.
4.Flash to 1000K and let it cool down to 550K under an O2 pressure of about 4E-8Torr . Repeat this step a few times (5 might be good).
5.Check for cleanness with a technique you have available (O2 TPD, Auger, etc.). You might have to repeat steps 3 and 4 a few times until the sample is clean.